What the ventilator actually controls.
The ventilator is a machine you can read. Five cards on what the screen is telling you: what a mode controls, reading the waveforms, asynchrony, trapped air, and the high-pressure alarm.
Reviewed October 2026 · verify against current guidelines
What the ventilator actually controls.
A mode is not a setting. It is three answers — what starts the breath, what the machine holds constant during it, and what ends it. Every mode name on the dial is one combination of those three.
The three questions — read a mode across the row, not by its name
| Mode | Triggered by | Holds constant | Ended by |
|---|---|---|---|
| Volume A/C | Patient or clock | Flow and volume | Volume delivered |
| Pressure A/C | Patient or clock | Pressure | The clock |
| PRVC | Patient or clock | Pressure, aiming at a volume | The clock |
| Pressure support | Patient only | Pressure | Flow falling to ~25% |
You can guarantee one, never both — compliance decides the other
| Volume control — volume is fixed | A stiffening lung is paid for in pressure. The tidal volume holds and the peak climbs, so the injury announces itself on the alarm. |
|---|---|
| Pressure control — pressure is fixed | A stiffening lung is paid for in volume. The pressure holds and the tidal volume falls away quietly, with no alarm unless you set one. |
The settings behind the mode
- Trigger sensitivity. About −1 to −2 cmH2O, or a flow trigger of 2–5 L/min. Too tight and the patient works to open the valve; too loose and the circuit self-triggers.
- Inspiratory time. Sets the I:E ratio, normally about 1:2. Shorten expiration and the lung does not finish emptying.
- Rise time and flow. How fast the breath is delivered. Too slow in a hungry patient looks like air hunger on the pressure trace.
Reading the waveforms.
Three traces, three questions. Pressure tells you what the breath cost. Flow tells you whether the lung finished emptying. Volume tells you whether what went in came back.
Two volume-control breaths — the second is held at end-inspiration
| Pressure | Peak — flow plus lung. Plateau — the lung alone. |
|---|---|
| Flow | Back to zero before the next breath. |
| Volume | What went in has come back out. |
Two holds, two answers — both need a passive patient
| Inspiratory hold → plateau | Two seconds with the breath held in. The gap between peak and plateau is airway resistance; the plateau itself is what the alveolus felt. |
|---|---|
| Expiratory hold → total PEEP | Three to five seconds at end-expiration. Anything above the PEEP you set is trapped air the lung could not release. |
The breath the patient didn't want.
Two pumps have to agree on when the breath starts, how big it is, and when it ends. Disagreement is common, it is visible on the traces before it is visible in the patient, and two of the four types injure the lung.
Four patterns on the traces — flow trace, except where noted
| Double trigger | Two breaths stacked, so the delivered volume doubles. Lengthen the inspiratory time or raise the set volume. |
|---|---|
| Ineffective effort | The patient asks and gets nothing. Usually trapped air or too tight a trigger. Fix the trapping first. |
| Flow starvation | On the pressure trace. Scooped out, because demand exceeds the set flow. Raise the flow or the pressure support. |
| Reverse trigger | The machine breath provokes the effort. Seen in deep sedation, and deepening it further makes it worse. |
Order of fixes
Settle the cause before the settings: pain, fever, delirium, trapped air, a blocked tube. Then the mode and the flow. Paralysis is the last option, not the first, and sedation alone will often make reverse triggering worse rather than better.
When the lung won't empty.
Expiration is passive, and it takes time. If the next breath arrives before the lung has finished emptying, the leftover volume stays — and its pressure never appears on the PEEP you set.
The flow trace tells you first — expiratory flow, below the zero line
| Emptied | Flow reaches zero and stays there. There is a flat gap before the next breath. |
|---|---|
| Trapping | The next breath cuts expiration off mid-flow. Every breath leaves a little more behind. |
Measure it, then unload it — an expiratory hold on a passive patient
| Total minus set | Auto-PEEP is the difference. Hold expiration 3 to 5 seconds and read what the pressure settles at. |
|---|---|
| Give it time | Lower the rate, shorten the inspiratory time, accept the rising CO2. Expiration needs the seconds more than the lung needs the breath. |
| Widen the exit | Bronchodilators, suction, the largest tube available. And treat the effort itself — a struggling patient traps more. |
When the pressure alarm sounds.
Take the patient off the ventilator and bag them. That single move separates the machine from the patient, tells you how hard the lung is to inflate, and buys the time to work through the rest.
Back on the ventilator, one question splits the differential — inspiratory hold: peak, then plateau
| Gap widens — resistance | Secretions, bronchospasm. A kinked or bitten tube. A clot in the lumen. Too narrow a tube. |
|---|---|
| Both rise — compliance | Pneumothorax, trapped air. Edema, consolidation. A distended abdomen. A stiff chest wall. |
The crashing patient — the four that kill, in the order you can exclude them
| Displaced | Tube out, or down a bronchus. Check depth, listen, confirm with capnography. |
|---|---|
| Obstructed | Secretions, a clot, a bitten tube. Pass a suction catheter — it either goes or it does not. |
| Pneumothorax | Absent sliding on ultrasound. Under tension it needs decompressing before imaging. |
| Equipment | Circuit, valve, water in the tubing. Bagging has already ruled this in or out. |
Educational reference only — not medical advice. Verify against local protocol.
Downloads